A gas chromatogram of the organic components of a sample of beer using a column that separates compounds on the basis of their relative boiling points provides a GLC trace with several peaks. Two of the smaller peaks, with retention times of 9.56 and 16.23 minutes, are believed to be ethyl acetate and ethyl butyrate, respectively. Suggest 2 ways in which you could adjust the conditions of the experiment so as to reduce the retention time for all components in the sample. |
|
Subject:
Re: Chemistry - GC & GCMS
Answered By: tehuti-ga on 01 May 2003 13:00 PDT Rated: |
Well met again! :) The retention time can be reduced for all components by using a shorter column and by increasing the temperature of the column. Source: The retention time (tR) is directly influenced by the column length. Longer columns cause higher retention times The column temperature influences the separation, too. An increased temperature induces a decrease of the distribution coefficient and the retention time. From Gas chromatography (GC) http://www.zal.tu-cottbus.de/zal/prakt/gaschrom.htm (web site of the Brandenburg Technical University, Cottbus, Germany) Search strategy: decrease "retention times" "gas chromatography" |